Most of the parts are the same for both instrument when it comes to the tracking generator, since it only goes up to 2.9 GHz. The only differences will probably be some of the cables.
There are some manual adjustments that would normally need to be made on the card cage TG / LODA control board when the option is added to an instrument, but since you would be moving all of the parts that the adjustments are for along with the control board the adjustments should be fine as they are.
You do not need to enable the option, the instrument will recognize the contol board and enable it all. The only adjustments that you should need to do is the front panel TG calibration that is run with a short cable connected between the RF OUT and RF IN connectors on the front panel and executed on the CAL menu.
Have fun,
Mike
|