The waveguide kits <26GHz (X,P,K11644A) do include adapters. None of the higher frequency kits include adapters - I am not sure why...
The WG cal kits contain hardware that can be used for either SOLT or TRL/TRM calibrations.
The kit definitions for 8720-series analyzers are set up for SOLT calibrations. There are reasons for this:
First, standard 8720 analyzers do not have the 4 receivers needed for a true TRL/LRM type calibration - they use "TRL*" which makes some assumptions that can compromise results. So, SOLT usually works better for a standard 8720 analyzer.
Second, 8720 analyzers have a firmware limitation - it is not practical to combine SOLT and TRL calibrations in a single definition. see here for more info:
http://na.tm.agilent.com/vnahelp/tip17.htmlNow, there was an option (option 400 and some other Hxx specials) that gave you the hardware to perform a "true" TRL calibration. If you have this option on your analyzer, you might want a different cal kit definition for TRL.
I don't have these set up for all the kits. I do have a TRL definition for the X11644A WR90 kit. See the .zip file attached to this message. Use this as an example. I don't have access to an 8722 here, so I can't set up the higher frequency kits for you. But it is not that big a job - you just need to rearrange the class assignments to make TRL work.
Note that the current PNA or ENA analyzers do not have either the hardware OR the firmware limitation.