Thread: E4402B Cal constants code needed


Permlink Replies: 1 - Pages: 1 - Last Post: May 5, 2010 12:09 PM Last Post By: tabbott
Steve Reeves

Posts: 23
Registered: 03/06/08
E4402B Cal constants code needed
Posted: Jul 29, 2009 3:43 PM
Click to report abuse...   Click to reply to this thread Reply
Can one of the Agilent gurus at the Spectrum Analyzer Division provide a list of IEEE commands for adjustment of the E4402B Spectrum Analyzer? I'm aware of the Agilent software platform for Performance Test/Adjustment but don't have the same lab standards that the s/w requires. Any help would be appreciated. I've done something similar with the 8591C/E with an automated Frequency Adjustment routine.

Steve Reeves
Cisco Systems
tabbott


Posts: 1,483
Registered: 07/24/09
Re: E4402B Cal constants code needed
Posted: May 5, 2010 12:09 PM   in response to: Steve Reeves in response to: Steve Reeves
Click to report abuse...   Click to reply to this thread Reply
Hi Steve -

Newer Agilent Spectrum Analyzers are much more complex than older analyzers (such as the 8591C/E). Therefore, adjustments are not a simple as they are for older analyzers. To calibrate / adjust and E4402B, you can either purchase the Agilent TME software (www.agilent.com/find/calibrationsoftware) or you can send in your instrument to an Agilent service center (www.agilent.com/find/assist)

Regards -