8561E Reference Level cannot be adjusted
Posted: Jun 7, 2010 12:40 PM
Click to report abuse...   Click to reply to this thread Reply
I have an 8561E that passes all diags and can display a consistent trace when swept through its full bandwidth. The problem is when I perform the Reference Level Calibration as shown in the User's Guide, Chpt 1the peak is too high.

I set the center to 300Mhz, Span to 20Mhz but the peak that is there is not -10dB, but slightly higher than 0dB. The Ref Level Adj between -528 and +528 barely varies the peak. Also the AmpCorr entries all seems to be empty. Is this possible a Calibration issue?

Thanks for any input.
Victor
Mike M


Posts: 223
Registered: 04/10/07
Re: 8561E Reference Level cannot be adjusted
Posted: Jun 8, 2010 12:46 PM   in response to: vjsilva1 in response to: vjsilva1
Click to report abuse...   Click to reply to this thread Reply
There is no Reference Level Offset turned on is there? If so, you will see an "R" on the left-center of the graticules.
vjsilva1

Posts: 20
Registered: 02/27/10
Re: 8561E Reference Level cannot be adjusted
Posted: Jun 8, 2010 2:11 PM   in response to: vjsilva1 in response to: vjsilva1
Click to report abuse...   Click to reply to this thread Reply
Thanks Mike. I believe I checked that because I did put in an offset of -11dB to get the Ref Level to the right level. After that I deleted the offset. I'll recheck again tonight.

How much of an adjustment should the Ref Level adjustment have from -528 to +528?

--Victor
Mike M


Posts: 223
Registered: 04/10/07
Re: 8561E Reference Level cannot be adjusted
Posted: Jun 8, 2010 2:32 PM   in response to: vjsilva1 in response to: vjsilva1
Click to report abuse...   Click to reply to this thread Reply
I get a 10 dB range on mine...
vjsilva1

Posts: 20
Registered: 02/27/10
Re: 8561E Reference Level cannot be adjusted
Posted: Jun 8, 2010 3:02 PM   in response to: vjsilva1 in response to: vjsilva1
Click to report abuse...   Click to reply to this thread Reply
I have maybe 1dB or 2dB max. Also it does not seem linear. For example if I am adjusting it up, it starts going up but then goes down again while I'm still increasing the level. Almost like a counter missing a high order bit that is being fed into a DAC. Maybe that could be the problem?

I did confirm I have no Amplitude Level Offset enable, no R on the left center.

I have checked the unit with an RF generator and it's clearly showing the input signal as too high across the whole bandwidth.
I input a sine at 300MHz, 1, 2,3,4GHz and they all show up about 10dBm too high. It's starting to look like there is a problem with the Ref Level adjust circuitry. I have the CLIP, any suggestions where to look for that circuitry?

--Victor
vjsilva1

Posts: 20
Registered: 02/27/10
Re: 8561E Reference Level cannot be adjusted
Posted: Jun 19, 2010 1:42 PM   in response to: vjsilva1 in response to: vjsilva1
Click to report abuse...   Click to reply to this thread Reply
Found the problem. It was related to the reference level circuit.
I traced it to the AF7837BN dual DAC not outputting the correct Ref Level.
Tracing back, all the control signals seemed fine except L_IA0, which comes from U505c (74HCT02).
This gate was bad and was preventing all writes to the RF Gain DACs.

--Victor