Dear Sir:

I need to measure a load accurately for a one port measurement. The problem is there are other devices interposed between my load and port 1 of my VNA such as adapters, fixturing connectors, etc.. I would like to use the NA's TDR function as a means to set gate start and stop times so that I may measure only my load in isolation.

Generally, the method to make this type is measurement is to set the gate start and stop times and then disable the transform function (although gating is enabled) and then view the load on a Smith Chart.

My question is this, in order to mark the best start and stop times using the gating function, what TDR function is best to use to "mark" the gate start and stop time.

I like to use the the impulse response function to mark the start gate time. I generally set the start time at the peak of the impulse response when I am certain I have detected the physical position of the load. Can you tell me, for the most accurate measurement measurement of my load should I set the start time at the base of the impulse response at the load point (the base point I am referring to marks the beginning of the upward slope of the impulse signal) or at the peak of the impulse response? Keep in mind the measurement I am making does not use the impulse response to make the measurement of the load itself, the impulse response is only used to mark the start gate time before I disable the TDR option (but gating is enabled).

To determine the optimum stop gate time to capture data referenced to only the load is the primary reason why I have written this post, can you tell me how I should go about determining the best gate stop time to make the most accurate measurement of my load?

Edited by: SOLT_guy on Aug 19, 2012 8:23 PM
Dr_joel


Posts: 2,369
Registered: 12/01/05
Re: VNA measurements and TDR gating
Posted: Aug 20, 2012 3:32 AM   in response to: SOLT_guy in response to: SOLT_guy
Click to report abuse...   Click to reply to this thread Reply
It's quite complicated to understand the proper gating effects for gating out a load. I explain much of this in Chapter 4 or my new book (it takes about 20 pages to fully explain gating and gating compensation).

The effects that must be included are:

Compensating for lost energy to to gating out connectors preceeding the DUT
Compensating for gate re-normalization due to non-symmetric gate responses
Compensating for gate shape effects

In practice, the last two effects occur only in the last 10% of the frequency response, if low-pass mode is used.

The gate start should be set at least one gate width before the onset of the DUT time-domain response. The gate stop should be set to center the DUT response in the gate. Wider gate widths generate smaller edge effects so make the gate as wide as practial while keeping the response centered on the DUT.

References: http://www.wiley.com/WileyCDA/WileyTitle/productCd-1119979552.html
SOLT_guy

Posts: 258
Registered: 05/30/09
Re: VNA measurements and TDR gating
Posted: Aug 20, 2012 3:01 PM   in response to: Dr_joel in response to: Dr_joel
Click to report abuse...   Click to reply to this thread Reply
Dear Dr. Dunsmore;

Thank you for your reply.

Since I have not yet "marked" a start or stope time, when you made this statement:

"The gate start should be set at least one gate width before the onset of the DUT time-domain response."

Can you please define the term "gate width"?

Thank you for feedback, it is appreciated.

Dr_joel


Posts: 2,369
Registered: 12/01/05
Re: VNA measurements and TDR gating
Posted: Aug 20, 2012 5:10 PM   in response to: SOLT_guy in response to: SOLT_guy
Click to report abuse...   Click to reply to this thread Reply
I forget the exact wording, but it is something like gate rise time. The gate has a rise time and 2x is the gate width, so you should be sure the gate start is before the DUT response by at least one rise time or probably at least 2 to ensure no effect from gate rise effects.
SOLT_guy

Posts: 258
Registered: 05/30/09
Re: VNA measurements and TDR gating
Posted: Aug 21, 2012 2:42 AM   in response to: Dr_joel in response to: Dr_joel
Click to report abuse...   Click to reply to this thread Reply
Dear Dr. Dunsmore;

Thank you for your reply.

Can you tell me if gate rise time is specified in the NA manual?

If "gate rise time" is not specificed can you please provide me with an equation, observed gate rise time by inspection, or an experimental set up to determine gate rise time?

Also, with regard to "stop time," for a one port measurement, where the load is the last element in the circuit, should I position the stop time at a time before the point in time when the second impulse signal arises (the second impulse signal is an undesired signal)?
SOLT_guy

Posts: 258
Registered: 05/30/09
Re: VNA measurements and TDR gating
Posted: Aug 25, 2012 7:53 AM   in response to: SOLT_guy in response to: SOLT_guy
Click to report abuse...   Click to reply to this thread Reply
Dear Dr. Dunsmore:

I would like to thank you for your past replies.

I bought your book.

The problem is your book ships out in about one month from today. I want you to know that I have not given up on researching this topic, I will be getting back to this thread.

I will try to e-mail you if I have any questions in the future in about one month from today.
Dr_joel


Posts: 2,369
Registered: 12/01/05
Re: VNA measurements and TDR gating
Posted: Aug 27, 2012 2:04 PM   in response to: SOLT_guy in response to: SOLT_guy
Click to report abuse...   Click to reply to this thread Reply
look at table 1.3 on page 29 of http://cp.literature.agilent.com/litweb/pdf/5989-5723EN.pdf

This gives the VNA rise time and I think the gate rise time is on the same order. The math doesn't require it but I think our implementaton uses the same values.

The stop time for the gate should be such that the desired response is in the center of the gated region.

P.S. I hope you enjoy the book, it has many details about topics that you have mentioned in the past.

Edited by: Dr_joel on Aug 27, 2012 1:04 PM